On 07/25/2017 06:57 PM, Nathan McCorkle wrote:
> for e.coli 1800 volts is the default across a 1 mm cuvette.
OK. Here are parameters you could use on separate loadings of the cuvette to see if
some survive better and if the volt threshold of poration is reached:
First load your cuvette with a suspension of cells in water, measure resistance with a DVM from one side of the cuvette
to the other aluminum metal contact side. It should be something like 1 or two mega Ohms. That's what I get with
2 grains of salt in a cuvette full.
Plug in the culture shock to USB, start your terminal emulator and see a prompt >>
Measured with Mitchell's culture shock with no load:
Simple program does pulses then stops, so exp up quick, exp down depends on load, no flat top.
1600V peak:
>> a(70,5,35)
>> pulse()
reload cuvette, save and mark that cell suspension [a(70,5,35) 1600V 2mm gap]
1800V peak:
>> a(70,5,37)
>> pulse()
reload cuvette, save and mark that cell suspension [a(70,5,37) 1800V 2mm gap]